Hello:
I am not sure I understand your problem: Molflow+ has been tested and validated extensively vs existing literature, in particular early papers on montecarlo simulations of transmission probabilities of tubes, like Smith and Lewin on Journal of Vacuum Science and Technology (see curve for the transmission probability vs L/R). This was the main subject of the 2009 paper, again on JVST A, by me and ESRF colleague Jean-Luc Pons.
You didn’t include a model in your message, but if I quickly make a simple model of 2 cubes connected by a pipe of constant cross section (circular or other shape) and I calculate with Molflow+ the conductance of the pipe via C=Q/dP, with dP being the pressure difference inside the two cubes (I inject molecules in one and I pump in the other cube, as one would do in a lab to measure the conductance), I get results which deviates from your formula, see line 3 (your formula 12.1.D^3/L) and line 2 (C=Q/dP). P2 and P80 in formula 2 are the average pressure on the bottom facet of the two cubes. Ratio of the 2 values is about 1.36.
You say, rightly, that C should change as 1/D^3, and this is exactly what Molflow+ calculates: for a given diameter D C changes as D/L, and then there is the D^2 factor for the inlet area, since the transmission probability depends only on D/L. I do not see any discrepancy.
The model I attach has two 20-cm side cubes, connected by a L=10 cm, D=2 cm round tube.Molflow+ calculates P_tr (Transm. Prob.) of 0.191 (rounded), as per Smith&Lewin, see intersection of the two red lines.
The “kinetic” formula for the conductance (see “Formula editor” screenshot) is given by C=A*11.7705*P_tr, where 11.7705 is the l/s of a 1 cm2 opening for mass 28 gas at 20 C (293.15 K), a value given by the Maxwell-Boltzmann distribution integrated over a 1 cm2 opening, the /4 factor, with the MB average molecular velocity. See formula 6 and 7. The Kinetic Factor has a factor of 1/40 in the formula to go from m/s in the MB formula and Pa.m3/s (SI units) to the “Molflow+ units” cm and mbar.l/s (factor of 1000 for m3 to l/s, divided by 100 to go from m to cm. times 4).
The zip file attached has all 6 facets in the pumping cube (on the right in the view, see screenshot) with sticking=1, i.e. all molecules exiting the tube are pumped, and cannot go back to the tube, but if you change the sticking of these facets, or mimicking a real lab system you set sticking>0 only on one of them (the one with the physical pump) you get a similar value for the tube conductanc from formula 2.
See how in the model the conductance calculated via C=Q/dP and the kinetic conductance via the transmission probability match quite well, formula 2 vs formula 5.
So, it is clear that the often used analytical equation for the conductance is not valid here, since it gives a value of 9.68 l/s vs 7.07 of the Formula 2 or 7.05 of Formula 5 (1.36x higher).
The reason is that the 12.1.D^3/L formula is valid for LONG tubes only, i.e. where the transmission probability in the Smith&Lewin figure gets its asymptotic value of 8R/3L, in my model we are at L/R=10, still on the “bending” part of the curve labeled “0”.
The factor 1.36 is the difference (in terms of transmission probability) of the intersection point of the inclined solid red line and the vertical dotted line, with respect to the intersection of the two dotted lines on the Smith&Lewin figure.
If I make the tube longer, e.g. 30 cm instead of 10, I get a ratio of the conductance values calculated via the approximate formula line 3 and that on line 2 much smaller, i.e. in better agreement with each other. I Include the zip file too, it has L30 in the file name, the other one having L10.
If I make the tube 100 cm long I get almost a perfect match for the values in the approximated formula and the C=Q/dP, formula 3 and 2 (ratio 1.044, see file with L100 in the name).
The conclusion is that the formula C=12.1.D`3/L is valid only in the limit of LONG tubes, i.e. when L/R > 50 (where the curve with label “0” in the figure of Smith&Lewin coincides with a straight line with inclination -1 on the log-log scale.
That’s all, if not clear or further questions do not hesitate to write back, will be happy to expand on this important issue of vacuum science and technology.
R.